Tektronix 2213

R

raymon

Jan 1, 1970
0
Does anybody have schematic of PSU/CRT section of Main Board. I
downloaded full service manual from BAMA site only to find this
schematic missing. Should be labelled No 9.
I have problem with scope recently purchased, low psu voltages
resulting in short horizontal trace.
 
A

Ancient_Hacker

Jan 1, 1970
0
raymon said:
Does anybody have schematic of PSU/CRT section of Main Board. I
downloaded full service manual from BAMA site only to find this
schematic missing. Should be labelled No 9.
I have problem with scope recently purchased, low psu voltages
resulting in short horizontal trace.

This is often a sign the filter caps are going bad. If there are four
chapstick-sized silver upright electrolytics near the inside corner of
the PS, they're probably the culprits.
 
R

raymon

Jan 1, 1970
0
Ancient_Hacker said:
This is often a sign the filter caps are going bad. If there are four
chapstick-sized silver upright electrolytics near the inside corner of
the PS, they're probably the culprits.

Thanks but have replaced these capacitors and PSU voltages still low
out of spec. Checked current limit resistor on pre-reg board and find
current is at limit value (>1A)
suggests one of the supplies is loading psu but cannot find which one
 
A

Andreas Tekman

Jan 1, 1970
0
raymon said:
...

Thanks but have replaced these capacitors and PSU voltages still low
out of spec. Checked current limit resistor on pre-reg board and find
current is at limit value (>1A)
suggests one of the supplies is loading psu but cannot find which one

The power supply of the 2213/2215 is a combination odf teh prereg ,
follwed by a swithcmode. The switchmode part runs from 42 VDC -- so
you can check the switchmode by external powering it from a Dc supply
(I recommend a lob type, 0-50V, 0-3A, CV/CC mode adjustable for
repair).

If the switchmode works well with the lab supply, go for the prereg:
the flywheel diode, a fas 400V/1A type near the prereg inductor caused
me a lot of headaches while looking for a fault similar to the one you
describe.


hth,
Andreas
 
J

Jim Yanik

Jan 1, 1970
0
The power supply of the 2213/2215 is a combination odf teh prereg ,
follwed by a swithcmode. The switchmode part runs from 42 VDC -- so
you can check the switchmode by external powering it from a Dc supply
(I recommend a lob type, 0-50V, 0-3A, CV/CC mode adjustable for
repair).

If the switchmode works well with the lab supply, go for the prereg:
the flywheel diode, a fas 400V/1A type near the prereg inductor caused
me a lot of headaches while looking for a fault similar to the one you
describe.


hth,
Andreas

I've been trading some email with him on this,and it appears his HV
multiplier may be the problem. Hopefully,not the CRT itself.
 
A

Andreas Tekman

Jan 1, 1970
0
Jim said:
I've been trading some email with him on this,and it appears his HV
multiplier may be the problem. Hopefully,not the CRT itself.

--

H ijim,

I had this once in a 2215, a dead HV multiplier. The smps part tried to
start up, but due to overloead it ended in continous "tweeeeeeet"
sound. E.g. swichting frequency turned too low, ended in audible range.

Desoldering the multiplier and everything worked well (except HV, of
course). The multiplier I later rebuild by discrete , 5 HV-diodes and 5
3.3 nf /3kv caps were handy, so this works well for years. It a x4
multiplier, and one (5th diode/cap) for the -2kV.

The original, molded HV-multiplier might be hard to get. So, the above
DIY solution is may be helpful.


hth,
andreas
 