my scope show test 04 fail 10
J'm looking for schematic for this scope ( J have manual without diagram
schematic )
probably scope need calibration or someting is damaged with memory
( sorry for my english )
Best regards from Poland
J
Jim Yanik
Jan 1, 1970
0
t.com:
schematic )
Bad EAROM perhaps.That means you may have lost your calibration constants
that the microprocessor loads into registers throughout the scope to set
gains,timing speeds,etc.
T
tekamn
Jan 1, 1970
0
Jim Yanik
jyanik
at
kua.net
04 fail 10 means bad checksum, bit 4 set. Or in other words: He has to
run through the cal process.
The OP can download the service manual for free from
www.teknetelectronics.com
This manual includes all the schematics and calibration process steps.
Good luck!
hth,
Andreas
R
rubens
Jan 1, 1970
0
Thank's for answers.
I will try to calibrate scope .
If it will not set then EAROM is bad.
Where is possible it to buy EAROM?
J
Jim Yanik
Jan 1, 1970
0
Good luck!
hth,
Andreas
If he recals,and then gets another fail message,then the EAROM is bad;it's
not retaining the data. We used to replace a lot of EAROMs at TEK while I