[email protected]
- Jan 1, 1970
- 0
Okay, sorry it took so long. I'm going to take your advice and try the
AD815 circuit. So, some answers to your questions...
Take a look at the three jpegs at
http://www.stanford.edu/~jwodin/lineartrap/LinearTrapPics/
The linear trap is mounted inside the vacuum system. The 136Ba+ ion
can be trapped in this linear trap anywhere between 1e-10 torr to 760
torr Helium (has to be very clean, so the system is pumped down with
the Pfeiffer turbo pump, as shown, and then baked for about a week at
150C). Residual gas partial-pressures (non-Helium) need to be below
1e-8 torr, or else the ion can capture on one of them, or be ejected
via a collision. As you can see, the ion is injected on the left hand
side, and is radially confined with the RF field. It is free to move
axially, which is why we need the DC potential (also shown) to bring
the ion down to the observation area, where it is viewed by exciting it
with 493nm and 650nm lasers and looking for it's fluorescence.
The vacuum feedthroughs that we are using are shown in the picture, but
you can see details at
http://www.mdcvacuum.com/urd/uniface.urd/ecf0070w.display?6.1.c.1
We use this because it is rated to UHV (ultra high vacuum ~1e-10 torr,
which means no finger grease allowed, all metal seal, etc.)
As far as actual electrical connections, my current thought is at
http://www.stanford.edu/~jwodin/lineartrap/LinearTrapPics/cutaway2.JPG
Each pin on a 10-pin feedthrough carries RF+DC for one pair of
electrodes. A kapton insulated wire
(http://www.mdcvacuum.com/urd/uniface.urd/ecf0070w.display?6.1.e.1)
connects the feedthrough pin to an electrode, and another short wire
jumpers that electrode to it's brother. Similarly, each pin on another
10-pin feedthrough carries DC for the other pair of electrodes, and is
connected in a similar way.
Note that I assume there will be heavy RF pickup on the electrodes that
are supposed to have only DC on them, so I was thinking of using your
circuit with the addition of 0.1uF caps (outside the vacuum system of
course, since caps are no good for vacuum). The naïve hope is that
any RF pickup on E2+E3 will go right through the 0.1uF cap (>> 25pF)
instead of re-radiating into the trap (so that the ion truly sees only
DC on those rods). I don't know if this will work.
.. | A1
.. |__________|\____ T1 ___________| E1+E4
.. | inv #||# 330 | (Set 1)
.. | | A2 #||# 10W |
.. | |___|\____| #___|________________________| E2+E3
.. | |/ | _|_ |
.. | AD815 | 1nF_|_ 0.1uF -,-
.. | dc1 -,- |
.. | |___| E1+E4 gnd
.. | ac sig 100k | |
.. | bus ,--/\/--' (Set 2)
.. | |____________________________| E2+E3
.. | | _|_ |
.. | dc2 0.1uF -,-
Gnd
The electronics (RF,DC) will be put into an enclosure very close to the
system (~1ft). I was going to use some mil-spec amphenol connectors on
the box (see upper picture at
http://www.stanford.edu/~jwodin/lineartrap/equipment/AmphenolPanelMount.pdf
, though with 10 pins) and then for the mating cable
connector,(http://www.insulatorseal.com/urd/uniface.urd/ecf0070w.display?9.1.2.2).
and to carry the signal to the feedthroughs on the vacuum system, I was
going to use some multi-conductor cable with grounded outer
foil-shielding, like Belden 9935
(http://www.stanford.edu/~jwodin/lineartrap/equipment/Belden9935.pdf).
I figure that at 1-10MHz, if my cables are a few feet, they don't
have to be waveguides, so non-coax should be fine. The foil sheath on
the Belden cable should be good enough to not make the cables antennas.
Whew, okay, that's a lot of info...
How does the trap and the vacuum system go together?
I'd like to know more detail about those connectors and feedthroughs.
.. | A1
.. | |/ | _|_ |
.. | dc1 -,- |
.. | | _|_ |
.. | dc2 0.1uF -,-
Gnd
Yes, but you still want to shield those lines,
transmitter/antenna city, you know, otherwise.