Reading Current of Solenoids

Omar Zeyad

Dec 18, 2015
6
Joined
Dec 18, 2015
Messages
6
Hi Guys,

I'm having some issues with a circuit that I've created to measure the current going through a solenoid.

Basically, I'm using AD8202 to measure the current being supplied to a load (solenoid),

Here is a small description of the circuit.
Capture.PNG

The expected current that will go through the solenoid is about 2 Amps! I already tested the circuit with a power resistor and I was able to read the current through PIN 5. but everytime I test the circuit with the real load (the solenoid). the AD8202 gets blows!!

What do you guys think? would this have to do with the back EMF from the solenoid? or is my design have an issue?

Thanks
 

Omar Zeyad

Dec 18, 2015
6
Joined
Dec 18, 2015
Messages
6
I was thinking of adding a diode to the solenoid (to drain the back EMF). but I was afraid that it will change the property of the circuit that I'm testing.

Any idea if there is another chip that can handle higher voltages?
 

duke37

Jan 9, 2011
5,364
Joined
Jan 9, 2011
Messages
5,364
Your circuit does not have the clamp diode so excessive voltage is generated on switch off.
Having a higher voltage chip may not work since the solenoid can doubtless produce several hundred volts if turned off fast enough.
 

Colin Mitchell

Aug 31, 2014
1,416
Joined
Aug 31, 2014
Messages
1,416
You need to have a completely separated power supply for the solenoid, so the back emf does not enter the chip.
 

davenn

Moderator
Sep 5, 2009
14,470
Joined
Sep 5, 2009
Messages
14,470
You need to have a completely separated power supply for the solenoid, so the back emf does not enter the chip.

his circuit and the datasheet circuit already show that fact Colin !!
a snubber diode is still required tho to stop the large back EMF voltages entering the IC sense input pins


D
 

GPG

Sep 18, 2015
452
Joined
Sep 18, 2015
Messages
452
The ground is shared so back emf is impressed on the sense terminals relative to com.
 

Omar Zeyad

Dec 18, 2015
6
Joined
Dec 18, 2015
Messages
6
Thank you guys, I'm going to try adding a clamp diode over the load.

I'm almost certain it's the back EMF, I tested the circuit with power resistors and didn't have an issue. but I did have the same problem when using solenoids.
 

Minder

Apr 24, 2015
3,561
Joined
Apr 24, 2015
Messages
3,561
I am wondering why you need to measure the current as if the solenoid is DC the current is predictable based on the resistance of the coil?.
If it is a case of confirming the current is actually present, then you do not need an A/D conv to do this?
M.
 

Omar Zeyad

Dec 18, 2015
6
Joined
Dec 18, 2015
Messages
6
I'm validating a circuit that powers up the solenoids (which are current controlled).
 

Minder

Apr 24, 2015
3,561
Joined
Apr 24, 2015
Messages
3,561
If I read that right then all you need is actual current confirmation, not the value itself!