Hi All
I am in the process of repairing a second HP1741A scope belonging to a local ham radio colleague. The first one was my own scope which now appears to be fine. The faults include non-working beam finder and a 50Hz non-sinusoidal ripple on the displayed trace with no input signal. Further investigation showed the +5v line was +3v and U2 and CR2 on the LV Power Supply PCB were much too hot to touch. This is a coincidence because U2 had failed on my own scope some months back and had to be replaced along with several of the bridge rectifiers - in this case a past owner had run the scope on 230v mains but with the mains tap set to 120V!
Does anyone know if this U2 issue is a known failure problem please and what may have caused it?
Once I get the PSU back into operation I will look at further tests and calibration.
I am in the process of repairing a second HP1741A scope belonging to a local ham radio colleague. The first one was my own scope which now appears to be fine. The faults include non-working beam finder and a 50Hz non-sinusoidal ripple on the displayed trace with no input signal. Further investigation showed the +5v line was +3v and U2 and CR2 on the LV Power Supply PCB were much too hot to touch. This is a coincidence because U2 had failed on my own scope some months back and had to be replaced along with several of the bridge rectifiers - in this case a past owner had run the scope on 230v mains but with the mains tap set to 120V!
Does anyone know if this U2 issue is a known failure problem please and what may have caused it?
Once I get the PSU back into operation I will look at further tests and calibration.